When I first started using the Accu Chek Spirit Combo, I always got bubbles in the cartridge but then I started thinking about going back to basics.
I didn't find the Accu Chek rep an awful lot of help as she demonstrated filling up cartridges by injecting air downwards and made it look really easy. What I didn't realise was that she demonstrated the technique of injecting 300u air downwards into a vial that had nothing in it :roll:
It was only when I left the hospital that I then became a bit angry with myself for not being able to pick up on what she was demonstrating.
The problem is I think in the design of the cartridges. The Animas cartridges are thin and only hold 200u insulin whereas the Accu Chek cartridges are wider and longer and can hold any amount upto 300u insulin. It’s therefore harder to fill the Accu Chek without any bubbles at all.
There are various ways to fill cartridges though and following the pump instruction manual doesn’t always help……
If people want to follow the technique that MPB has mentioned then they need to be a bit careful when injecting air downwards into a vial that is full up (new) with insulin when using Accu Chek cartridges. If 300u of air is injected downwards into a full vial, then because there is nowhere for the air to escape to, the plunger will rise before you can turn the insulin vial upside down to withdraw. I once lost nearly half a vial of insulin as the plunger in the cartridge came out of it and the seals blew off. I would recommend therefore that when injecting air downwards into a full vial for the first time that the first cartridge contains no more than 200u insulin then they wont have the problem I had. The next fill up will be ok though as the insulin in the vial will be reduced so there’s more room for the air and you can fill up with 300u ……..
I still get my cartridges upto body temperature (boob temp) for about 10hrs regardless of whether I use room temp vial or one that has been in the fridge, as I keep my pump inside my bra. By doing so it makes air bubbles easily rise and then I just sharp flick with my thumb nail against the side of the cartridge to make the bubbles join up to being just one and then tilt cartridge slightly to get the bubble in the centre of cartridge at the bottom of the neck and then sharp flick a couple of times and then screw in plunger and push the bubble out. This works fairly well for me. I then put push on the black adaptor with a new tube already screwed in and put the cartridge in the pump and then let the pump go through the prime etc……….
Hope that helps some of you using Accu Chek