Hi Dennis,
Thanks again for your replies. That was very helpful. Thanks also fro mentioning the freebie control solutions, accu-check are now sending me some of these. Funnily enough, I had a call back from the accu-check careline tonight, the person I spoke to earlier in the day was a trainee apparently and hadn't got the information exactly right. The person who called me back gave me a lot more detail and some of it is quite useful info, so I thought I would copy what he said here in case anyone else is interested:
Accu-Check Aviva Meter Careline made the following statements re meter accuracy (3rd Feb ’09)
The MHRA allow a margin of error of 15% across the standard scale from 2.8-33.6
At the lower end of the scale, which runs from 0.6-2.7 the allowable margin is an absolute value not a percentage. The figure is 0.8 mmol/l (a margin of error of more than 100% on 0.6 and of 28% at 2.7).
The nature of variations in an individual meter is difficult to predict, there is some expectation that they would be more likely to be in one direction only, but they could also vary in both directions.
Whilst the control solutions give a range for both Level 1 (low BG) and Level 2 (high BG) controls, this is to take into account deterioration of both test strips and control solutions after their respective containers have been opened.
If both test strips and control solutions come from newly opened bottles (opened immediately before you calibrate your meter) then, the value of each control solution in mmol/l should be exactly in the middle of the relevant range.
The above point means that, by using newly opened test strips and control solutions, you can check the specific ‘margin of error’ for your own particular meter. Testing it several times over a fairly short period (so that the strips and controls are still fresh) will give an idea of whether there is any ‘random’ variation (if there is a lot of random variation its probably best just to get rid of the meter).
8)