Calibration of glucose meters are hit and miss.
From this website, it states.
I
SO guidelines state that blood glucose meters should provide results which fall between the upper and lower error bounds, given in the table below, 95% of the time:
- Within ± 0.83 mmol/L of laboratory results at concentrations of under 4.2 mmol/L
- Within ± 20% of laboratory results at concentrations of 4.2 mmol/L or more
Which is quite a reasonable figure, considering you have the basic meter accuracy, which is going to be a percentage, plus an offset, (eg +/- 5%, +/- 0.1 mmol/L for example) then on top you have the variability of the test strips, which are manufactured to another tolerance, on top of the meter tolerance. So they could be +10% in one batch, and -10% in the next batch, with the meter accuracy on top of that.
(I have made up the figures here, apart from the ISO guidelines quoted - could be any combination)
It would be reasonable to assume the meter should be fairly consistent from one reading to another, but needs to be kept at room temperature, as it's going to filled full of cheap electronics.
Test strips of better made now too, (no need to code mine anymore, batch to batch variance must be negligible, but only within the accumulated tolerance)
But you could still get the swing from the meter+the test strips in one batch, to the meter-the test strips in the next.
One thing that springs to mind as well is blood temperature.
I know it's supposed to be constant, but not when it's on your fingertips. Mine can be freezing at times, and the blood must be cooled.
We should try a few as a group, cold finger, warm up our hands, try it hotter, and see if there is any change.
Should still be in range though.
edit - just checked the spec on my meter.
Nexus meter system Operating condition: 10°C to 40°C (50°F to 104°F), below 85% R.H. (no condensing). The Meter, Test strips and Control solution should all be at room temperature (20°C-25°C / 68°F-77°F) before testing.
Not a great range to test in..