Hi
@bizzilizzi
The Freestyle Libre is great for trending but less so for absolute blood glucose. Most of us seem to find that it's reasonably accurate between say 4 to 9 mmol but outside of this range the magnitude of it's errors increase the further away from 'normal' one gets.
I don't know what meter you are using but blood test meters (sold in the UK) must comply with the requirement to be accurate within +/- 15% for 95% of readings.
https://www.diabetes.co.uk/blood-glucose-meters/blood-glucose-meter-accuracy.html
As I understand the technology, the biggest problem isn't the meter itself, but the enzymes in the test strips which can be influenced by temperature and air time among other things.
You then have to consider other factors like levels rising or falling sharply, the fact that your glucose levels are constantly changing and that your blood is not completely homogenous. The glucose in blood near your digestive organs starts to rise before the blood in your finger tips, and then has to be pumped round everywhere else.
You can get special solutions from the manufacture (usually FOC) to test your meter with. Also worth trying a different pot of test strips.
For what it's worth I had a similar experience at lunchtime today with 6 readings in the range 4 to 7.9 mmol in a couple of minutes, from 1 finger prick.
Normally you should get better results than what you're reporting above.
Edited to add, if you type libre in the search bar top right, you will find many threads on the subject.