Hi, My name is Ian, I have poorly regulated type 2 diabetes due to the fact that I'm bed bound 24/7.
I have two test meters, both obtained earlier this year
An Accu-chek mobile and a one touch ultra easy.
The confusion arises as the meters give completely different readings, for instance a few mins ago I tested my blood on both machines, the accu-chek gave a reading of 17.3 mmol, the one touch gave a reading of 12.2 mmol.
This is a large difference and I really don't know which meter to believe at present.
Any thoughts or ideas?
Accu check send you the calibration fluid for freeIt is worth ordering some calibration fluid. and checking your meters against it for your own peace of mind.
From testing a few different meters I found they rarely agree and the readings have a wider spread in the higher end of the scale.
@Administrator Could I suggest that someone adds a Sticky about the ISO accuracies, and include ISO (2016) due soon. Perhaps a table showing how the percentages convert to +/- mmol/L readings. Also, it appears that some meters are calibrated for Plasma, and others are Whole Blood, which is I believe a factor of 1.2 A list of meters in each category might help here. And lastly, a section showing how to convert USA readings (mg/dl) to mmol/L may be of interest to those of us reading research papers on the netAlthough ISO standards are changing, current meters only have to be +\-20% accurate with high glucose levels.. So you likely have 1 reading low and the other reading high.
If your true value was say, 15, your error could be 3 (20%) either way. Hope that makes sense.
Choose the meter that you feel most comfortable with and stick with it.
@Administrator Could I suggest that someone adds a Sticky about the ISO accuracies, and include ISO (2016) due soon. Perhaps a table showing how the percentages convert to +/- mmol/L readings. Also, it appears that some meters are calibrated for Plasma, and others are Whole Blood, which is I believe a factor of 1.2 A list of meters in each category might help here. And lastly, a section showing how to convert USA readings (mg/dl) to mmol/L may be of interest to those of us reading research papers on the net
PS the % to mmol conversion I suggested is for showing how a +/= 20% error looks like at different readings, i.e. at 3, 5, 10, 15 and 20 mmol/L actual bgl. It is not so bad at 20%, but soon it will be 15% which needs a calculator. I think many people on this forum are unaware of how far the readings can be out and still be within spec. This is probably more important for those on insulin/All I was suggesting is that as this subthread covers blood monitoring, then it was the first place I would tend to look, but I don't want to wade through pages of posts to find this basic info. I suggested it as a sticky so it is easy to find, and yes, maybe it is just a set of links as you have provided here. The subject of whole blood vs plasma has been raised in several threads, and not just by me, and there seems to be a lack of knowledge in the responses. I think that it becomes a whole blood calibration issue where a meter is designed to allow alternate site testing, but it is not made clear, and also the BGL range data limit info published for Newbies here on this forum does not acknowledge that some meters will read high by a factorof 1.2 I suspect my SD Codefree is whole blood since it consistently reads high, andI think my old XCEED meter suffered the same problem (both my XCEEDs read about 3 mmol high.) There is nothing in the literature that mentions whole blood, but I think all the ones I have mentioned can do alternate site tests. When I compare my SD against my Hba1c, then on both occasions it read around 1.4 mmol higher than the plasma equivalent on two occasions now.
I have finally found the relevant section where there is a Table showing the 2012 ISO accuracies vs readings as I suggested above. The link is http://www.diabetes.co.uk/blood-glucose-meters/blood-glucose-meter-accuracy.html@Administrator Could I suggest that someone adds a Sticky about the ISO accuracies, and include ISO (2016) due soon. Perhaps a table showing how the percentages convert to +/- mmol/L readings. Also, it appears that some meters are calibrated for Plasma, and others are Whole Blood, which is I believe a factor of 1.2 A list of meters in each category might help here. And lastly, a section showing how to convert USA readings (mg/dl) to mmol/L may be of interest to those of us reading research papers on the net
I finally found the ISO (2013) Table showing Error vs readings in mmol/l as requested above at http://www.diabetes.co.uk/blood-glucose-meters/blood-glucose-meter-accuracy.html@Administrator Could I suggest that someone adds a Sticky about the ISO accuracies, and include ISO (2016) due soon. Perhaps a table showing how the percentages convert to +/- mmol/L readings. Also, it appears that some meters are calibrated for Plasma, and others are Whole Blood, which is I believe a factor of 1.2 A list of meters in each category might help here. And lastly, a section showing how to convert USA readings (mg/dl) to mmol/L may be of interest to those of us reading research papers on the net
Although I completely agree with this the fact that certain meters read consistently high or low does make a difference if on insulin and are a driver. We are all aware of the DVLA guidelines and how strict they are ( as they should be ) so a meter that read lower than it should can make a difference.I come back to a point I made elsewhere on this topic though. Diabetes is not a condition of exactitudes. So one meter reads differently from another? So what? As a T1, you had to use urine and test tubes and then visual colours to guesstimate what was going on. Now you get an electrochemical reading that, due to reading to 0.1 of a mmol, people believe to be "accurate".
Given that timing, absorption rate, test locations and measurements are all variable and not at all "accurate" then the numbers on blood glucose meters can only ever be a guide. The best you can hope for is consistency. Getting hung up on the accuracy of the numbers or even the accuracy of the numbers isn't going to dramatically change your life. If you target 5mmol/l on a blood glucose meter, then, based on the current and future ISO requirements, you will be measuring within the range you want. with it is 4, 5 or 6 really doesn't matter a vast amount. For a type 1, what's more important at that point is the direction of travel.
This is why there's the five to drive guidance, and equally, if your meter shows you at 4.8, the DVLA guidelines say you can still drive. Even if the meter is reading 20% high, and your number is 4.5, with a real number that is 3.6, your meter says 4.5 (so no issue on that point) and you aren't technically or clinically in a hypo state at 3.6!Although I completely agree with this the fact that certain meters read consistently high or low does make a difference if on insulin and are a driver. We are all aware of the DVLA guidelines and how strict they are ( as they should be ) so a meter that read lower than it should can make a difference.
I too look for trends though rather than worry too much about single readings and hopefully with technology moving so fast, the margins of error will reduce even further in the future.
If he is in the UK, then this shouldn't be the case, especially on a pump. I've tagged @noblehead who can provide a letter for him to take to his GP to resolve that situation.He is limited on strip allocation, so cannot afford to do much retesting.
If he is in the UK, then this shouldn't be the case, especially on a pump. I've tagged @noblehead who can provide a letter for him to take to his GP to resolve that situation.