Well-Known Member
Background - T1 for over 55 years
Because of changes to the required accuracy level OneTouch notified users of the coming withdrawal of the Ultra range of meters.
As a result of this, after testing various makes of meter I decided to use the OneTouch Verio range.
I very quickly started to disbelieve the reported results based on how I felt.
OneTouch have been brilliantly helpful - - - BUT
when using Control solution to verify results on old and new meters (3 old and 3new) the following became obvious.
Each type of meter produced the same results for the respective control solutions which confirmed consistent operation of each type.
The old meters all displayed in the lowest 15% of the permissible range.
The new meters all displayed in the TOP 10% of the permissible range.
For me this translates into a discrepancy to the results which I have become accustomed over about 20 years
The comparison readings obtained by testing the same blood drop on 2 meters shows a position as below
Ref...............New..........Old.............Diff..........Diff as %age of Old
A...................3.4............2.8.............0.6.................21%
B.................10.3............8.4.............1.9.................22%
C.................16.2..........12.2.............4.0.................33%
I am aware that the rules on accuracy of meters have been updated .
I have 2 annoyances -
My brain is struggling to deal with the change in readings and allow me to act appropriately.
People who tell me not to compare old and new meter readings . . . . . ... . .
Has anyone else spotted this ??
Because of changes to the required accuracy level OneTouch notified users of the coming withdrawal of the Ultra range of meters.
As a result of this, after testing various makes of meter I decided to use the OneTouch Verio range.
I very quickly started to disbelieve the reported results based on how I felt.
OneTouch have been brilliantly helpful - - - BUT
when using Control solution to verify results on old and new meters (3 old and 3new) the following became obvious.
Each type of meter produced the same results for the respective control solutions which confirmed consistent operation of each type.
The old meters all displayed in the lowest 15% of the permissible range.
The new meters all displayed in the TOP 10% of the permissible range.
For me this translates into a discrepancy to the results which I have become accustomed over about 20 years
The comparison readings obtained by testing the same blood drop on 2 meters shows a position as below
Ref...............New..........Old.............Diff..........Diff as %age of Old
A...................3.4............2.8.............0.6.................21%
B.................10.3............8.4.............1.9.................22%
C.................16.2..........12.2.............4.0.................33%
I am aware that the rules on accuracy of meters have been updated .
I have 2 annoyances -
My brain is struggling to deal with the change in readings and allow me to act appropriately.
People who tell me not to compare old and new meter readings . . . . . ... . .
Has anyone else spotted this ??