Apologies for slightly confusing matters by saying I don't use libre, it's the official app I don't use.
The confusion I think, is that some people think the libre extrapolates forward to give an estimate of current blood glucose (derived from current ISF, which lags actual blood glucose.) This is what I was asking about, but I agree that the regulators probably wouldn't be too happy with a device that is providing an extrapolated estimate rather than something based on the current ISF value. Nevertheless such an extrapolation method if it could be made to work would be an interesting thing.
I didn't explain myself very well, the value of interest to most users is actual blood glucose rather than ISF glucose level, so when one comes from doing finger prick testing and knowing what the value is now (well sort of as I guess there's also a slight delay in peripheral vs central glucose level responses), the display is telling you your "blood glucose" is such and such, but that's actually what it was 5 min (or however long) ago. I realise it's not actually blood glucose being measured, but the algorithms are giving a value which isn't direct ISF glucose concentration but rather an estimate of what actual blood glucose would be to generate a given ISF glucose concentration. Or is there no concentration difference (aside from the time-dependence) so there is no ISF to blood step required?Sorry I've no idea what you mean by time offset and working in the past?
Importantly, no CGM will attempt to tell you what your BG is as it has no access to blood. Medical device regulators would not allow a device that produces actionable information that does not reflect the source of the data. All CGMs report only on ISF.
The confusion I think, is that some people think the libre extrapolates forward to give an estimate of current blood glucose (derived from current ISF, which lags actual blood glucose.) This is what I was asking about, but I agree that the regulators probably wouldn't be too happy with a device that is providing an extrapolated estimate rather than something based on the current ISF value. Nevertheless such an extrapolation method if it could be made to work would be an interesting thing.
That sounds like you think it's relatively constant for a given person-sensor position combination - I was curious as to whether it might change with temperature, hydration and heart rate (so basically things that change peripheral blood flow and blood turn-over in general.)That's a good question - how long the lag? I've seen 2-45 mins depending on where you look and it can depend on where the CGM sensor is placed but mostly on the individual. The lag is due to movement of glucose from blood to ISF so if you have circulatory or cardio-vascular issues, ion imbalances, an infection or other illnesse. the lag may change.
Is it that the lag is minimal or the effect of any lag is lower with a smaller delta? I'm guessing it's the latter unless there's a concentration gradient effect at play?What is always the case is that lag time is minimal when glucose levels are relatively consistent. It's when they change over a short period of time that there is a noticable difference between BG and ISF levels.