Last month I switched, again, from my Roche Accu-chek aviva meter, to my CareSens-N meter. (Change of country = change of which meter and test strips are subsidized.) I immediately noticed higher than expected readings. As I had had a bad experience previously, with inaccurate readings from my CareSens-N meter, I did a dual reading (using my very last accu-chek test strip - last available, and last subsidized one at least!). Accu-chk = 5.5, CareSens = 6.6. This is the same difference that I had been noticed before, a year or so previously, checked at a medical centre in Sweden, against that practice's very reliable BG meter. (They also had an HBA1c reading device. The Accu-chek readings matched the HBA1c readings, over time.) (When I had used the CareSens meter, my BG readings and the subsequent HBA1c did not match up.)
Aotearoa/NZ changed over from the Accu-chek meter to the CareSens maybe in 2013-2014? As the subsidized meter of choice. Many diabetics were most upset (according to some feedback postings found online), as they enjoyed a high level of security and customer support with the Roche accu-chek meters. Especially, from T1 diabetics, as one can imagine.
The practice nurse at my doctor's office in NZ and I did an experiment. I tested my CareSens-N, and she tested with the practice CareSens-N, and I had my blood tested as part of a blood lipids test at a lab within 15 mins.
My Care-Sens = 7.2
Nurse's Care-Sens = 8.1 (same droplet of blood)
Lab's spot BG reading = 6.5
Additionally, for a month (previous to this post) I made adjustments (of approx 1.1 point lower) from my CareSens reading to get what I thought would be a more accurate reading, and predicted my HBA1c on that prediction - which turned out to be spot-on correct.
Anybody else have experiences of the CareSens meters - good or bad or neutral?
Aotearoa/NZ changed over from the Accu-chek meter to the CareSens maybe in 2013-2014? As the subsidized meter of choice. Many diabetics were most upset (according to some feedback postings found online), as they enjoyed a high level of security and customer support with the Roche accu-chek meters. Especially, from T1 diabetics, as one can imagine.
The practice nurse at my doctor's office in NZ and I did an experiment. I tested my CareSens-N, and she tested with the practice CareSens-N, and I had my blood tested as part of a blood lipids test at a lab within 15 mins.
My Care-Sens = 7.2
Nurse's Care-Sens = 8.1 (same droplet of blood)
Lab's spot BG reading = 6.5
Additionally, for a month (previous to this post) I made adjustments (of approx 1.1 point lower) from my CareSens reading to get what I thought would be a more accurate reading, and predicted my HBA1c on that prediction - which turned out to be spot-on correct.
Anybody else have experiences of the CareSens meters - good or bad or neutral?