Distributed by Menarini Diagnostics UK - if test strips have been exposed to high humidity, the results can be falsely high (MDA/2015/039)
GlucoMen LX PLUS meter
Read more: https://www.gov.uk/drug-device-aler...-blood-glucose-levels-risk-of-over-estimation
- Ensure that patients, or their carers, are able to use this device and that they are aware of what to do if a result is out of their normal range
- To avoid false high readings, advise them to follow the instructions for use, which includes making sure that the strips are not exposed to high environmental humidity for a prolonged period of time. Users should always:
- close the vial immediately after each use
- store the test strips in their original vial
- write the discard date on the vial label, which should be 9 months after the date of first opening the vial itself