Hello to my friends across the pond - and all over the world!

Just thought I'd share my recent experiences - maybe it will help some folks?

I am 62 years old -and a T2 (Highly Insulin Resistant) diabetic - using Medtronic pumps for about 20 Years. I use 5x concentrated U500 insulin (Humulin R) - which is considered 'off label' and not officially approved for use with Medtronic pumps here in the USA. But it has worked well in all manual mode pumps for years.

It also worked 'reasonably' well in the 780G - with some 'quirks' and annoyances (Some overnight and early morning lows - not due to over bolusing or other user error).

I was upset that SmartGuard was not throttling my basal rates to assist in preventing these lows. Originally I thought U500 was the sole culprit - after all it is a rather wonky & unusual insulin with unique Onset, Peak and Durations. I convinced my Doctor to go back to U100 (Novolog) - as a test - even though that would mean near daily replacements of reservoirs.

After originally training Smartguard on U500 rates - it was initially a challenge getting the pump to 'relearn' my new basal pattern rates which were now approximately 5 times higher using a U100 Insulin.

Working with Medtronic Tech support was a huge disappointment. Like talking with a robot - reading from a pre-pepared troubleshooting script. An offshore Phillipino - whom was friendly - but not familiar with diabetes, their pump, the mysterious SmartGuard or it seems anything else - near useless - with no further escalation path.

No one over there had the answers - no one had ever asked them how to put pump back into Smartguard Warmup and learning mode. They ended up sending me a refurbished pump (which I didn't use) as the the real answer came from a Medtronic Clinical Manager assigned to my Doctor for new user training. The answer was simple. Turn off SmartGuard and run the new Basal patterns in Manual Mode for at least 7 days. Turn SmartGuard back on - and voila - it learned my new rates.

Without going into detail - the pumps SmartGuard 'algorithm' - IMO - worked better using a Rapid acting U100 Insulin like Novolog.

This pump was doing a marvelous job at keeping me 'In Range' - better than I could do myself with manual mode pumps. But was it is really 'In Range'?

The real failure of this system (here it comes - NO suprise to many) - was the Guardian 4 sensors.

Worn properly on underside of upper arm - these sensors were consistently off by at least 30 mg/dl - and usually off by more than 50 mg/dl. Further, they seemed unable to measure and detect in my low ranges. I would be in critical low (<55 mg/dl) blood sugar (as measured by both Dexcom and by finger sticks and symptomatically) - yet Guardian G4 reported normal blood sugars values of 100 mg/dl or more. Calibrations were never effective - and they seemed ignored (no matter when they were performed).

So SmartGuard was managing me to a target of 100 (when in actuality it was a target of 150). Further, SmartGuard never saw low glucose trends - and as such, it never reduced or suspended insulin delivery - further contributing to my hypoglycemic events.

I have to wear two CGM's - Medtronic's to keep SmartGuard running - and Dexcom G6 to properly measure blood sugar and detect my lows. This is not sustainable.

I have decided to abandon therapy using this marvelous pump - due to the inaccurate and unacceptable G4 sensors.

Perhaps someday I will return - after they release their new sensor in the USA - or they wisen up - and start supporting sensors other than their own. G3 & G4 Sensors were both fails - will they ever get it right?

On the path to try a Tandem tslim X2 at my own expense (Insurance company won't pay for another pump for 4 more years). While the Tandem pump may have an 'inferior' algorithm/features to Medtronic's 780G - atl east it will be managing my rates based on true blood sugar levels - using a tried and true technology like Dexcom G6 (I like G6 better than G7). Good luck everybody!

Doug
(Melville, New York, USA)
 
Last edited:

EllieM

Moderator
Staff Member
Moderator
Messages
9,786
Type of diabetes
Type 1
Treatment type
Pump
Dislikes
hypos and forum bugs
Thanks for sharing your story and welcome to the forums @Mr. Freeze . It's really interesting to hear from a pumper who is also very insulin resistant.

I know some forumites claim that a new sensor will be available for medtronic later this year ? There are certainly plenty of complaints about the guardian sensors, along with happy users of the pump itself. Maybe keep the medtronic as a back up in case a new sensor comes out before it's out of warranty?

I had to choose between the tandem and the medtronic earlier this year and chose the tandem purely because of the sensors...

Once more welcome.
 

Antje77

Guru
Retired Moderator
Messages
20,268
Type of diabetes
LADA
Treatment type
Insulin
I'm afraid you'll have a hard time finding T2's on a pump on this predominantly UK based forum, let alone anyone using U500 off label in their pump.

Still, I find your experiments very fascinating and I hope you'll keep us updated on your findings!

I'm a T1 with relatively high insulin needs, so even if I'd want a pump, it likely wouldn't last for 3 days for me. Happy on pens at the moment, but should this ever change in the future, your experiences may be useful to me. :)

I'll include this chart for you, it nicely converts US/UK systems for hba1c from % (US) to mmol/mol (UK), and also blood glucose from mg/dl (US) to mmol/l (UK), very useful when reading internationally.

1719351219762.png
 
  • Like
Reactions: Prancy

Mr. Freeze

Member
Messages
7
Type of diabetes
Type 2
Treatment type
Pump
Thanks Antje77 and EllieM.
After posting this content I realized that many users on this forum - would have to convert my 'US measurement values' - and may not be familiar with the brand names of Insulins used here in the USA. Thanks again - it's great to be here!
 

EllieM

Moderator
Staff Member
Moderator
Messages
9,786
Type of diabetes
Type 1
Treatment type
Pump
Dislikes
hypos and forum bugs
After posting this content I realized that many users on this forum - would have to convert my 'US measurement values' - and may not be familiar with the brand names of Insulins used here in the USA. Thanks again - it's great to be here!
If you want an exact figure (and you are quite entitled to use US measurements), and you are good at mental arithmentic, .you just divide by 18 to do the conversion from mg/dL to mmol/L.
 

Prancy

Well-Known Member
Messages
383
Type of diabetes
Type 1
Treatment type
Pump
Hello to my friends across the pond - and all over the world!

Just thought I'd share my recent experiences - maybe it will help some folks?

I am 62 years old -and a T2 (Highly Insulin Resistant) diabetic - using Medtronic pumps for about 20 Years. I use 5x concentrated U500 insulin (Humulin R) - which is considered 'off label' and not officially approved for use with Medtronic pumps here in the USA. But it has worked well in all manual mode pumps for years.

It also worked 'reasonably' well in the 780G - with some 'quirks' and annoyances (Some overnight and early morning lows - not due to over bolusing or other user error).

I was upset that SmartGuard was not throttling my basal rates to assist in preventing these lows. Originally I thought U500 was the sole culprit - after all it is a rather wonky & unusual insulin with unique Onset, Peak and Durations. I convinced my Doctor to go back to U100 (Novolog) - as a test - even though that would mean near daily replacements of reservoirs.

After originally training Smartguard on U500 rates - it was initially a challenge getting the pump to 'relearn' my new basal pattern rates which were now approximately 5 times higher using a U100 Insulin.

Working with Medtronic Tech support was a huge disappointment. Like talking with a robot - reading from a pre-pepared troubleshooting script. An offshore Phillipino - whom was friendly - but not familiar with diabetes, their pump, the mysterious SmartGuard or it seems anything else - near useless - with no further escalation path.

No one over there had the answers - no one had ever asked them how to put pump back into Smartguard Warmup and learning mode. They ended up sending me a refurbished pump (which I didn't use) as the the real answer came from a Medtronic Clinical Manager assigned to my Doctor for new user training. The answer was simple. Turn off SmartGuard and run the new Basal patterns in Manual Mode for at least 7 days. Turn SmartGuard back on - and voila - it learned my new rates.

Without going into detail - the pumps SmartGuard 'algorithm' - IMO - worked better using a Rapid acting U100 Insulin like Novolog.

This pump was doing a marvelous job at keeping me 'In Range' - better than I could do myself with manual mode pumps. But was it is really 'In Range'?

The real failure of this system (here it comes - NO suprise to many) - was the Guardian 4 sensors.

Worn properly on underside of upper arm - these sensors were consistently off by at least 30 mg/dl - and usually off by more than 50 mg/dl. Further, they seemed unable to measure and detect in my low ranges. I would be in critical low (<55 mg/dl) blood sugar (as measured by both Dexcom and by finger sticks and symptomatically) - yet Guardian G4 reported normal blood sugars values of 100 mg/dl or more. Calibrations were never effective - and they seemed ignored (no matter when they were performed).

So SmartGuard was managing me to a target of 100 (when in actuality it was a target of 150). Further, SmartGuard never saw low glucose trends - and as such, it never reduced or suspended insulin delivery - further contributing to my hypoglycemic events.

I have to wear two CGM's - Medtronic's to keep SmartGuard running - and Dexcom G6 to properly measure blood sugar and detect my lows. This is not sustainable.

I have decided to abandon therapy using this marvelous pump - due to the inaccurate and unacceptable G4 sensors.

Perhaps someday I will return - after they release their new sensor in the USA - or they wisen up - and start supporting sensors other than their own. G3 & G4 Sensors were both fails - will they ever get it right?

On the path to try a Tandem tslim X2 at my own expense (Insurance company won't pay for another pump for 4 more years). While the Tandem pump may have an 'inferior' algorithm/features to Medtronic's 780G - atl east it will be managing my rates based on true blood sugar levels - using a tried and true technology like Dexcom G6 (I like G6 better than G7). Good luck everybody!

Doug
(Melville, New York, USA)
Mr. Freeze, I’m so sorry you had such a had experience. I wish there was a way to salvage your 780 treatment. I had a rocky start, but eventually got into zone. It’s now amazing. I wish someone could help analyze the issues and sort them out with solutions.

My pump now is almost always the same as my finger stick meter. I found that It works better if I calibrate, even though I don’t have to. I calibrate twice a day. I get such good control with Novolog. I am so grateful and I give thanks multiple times a day. I hope you’ll give 780 another chance sometime soon. Nothing against other pumps, but they all have issues from what I’ve read and yes I used Dexcom years ago, but the 780 has changed my life.

I’m attaching 2 of my data pages from the last 48 hours of my progress. Click on it to see info. I’ve been 100% TIR with 111-112 average BG. I wish everyone could have access and it worked for them. Best of luck to you.
 

Attachments

  • 90575D48-BD08-4F9F-8588-83CECEC4C941.png
    90575D48-BD08-4F9F-8588-83CECEC4C941.png
    273.1 KB · Views: 70
  • A461D571-4305-484C-83DD-BB6203A93D8D.png
    A461D571-4305-484C-83DD-BB6203A93D8D.png
    271.3 KB · Views: 66

Mr. Freeze

Member
Messages
7
Type of diabetes
Type 2
Treatment type
Pump
Hi @Prancy

Your TIR is fabulous! I am hopeful the Tandem TSLIM x2 will keep me in range too - like the 780G did.

Yes, I am disappointed to have to put this $5000+ USD pump in the drawer like an expensive paperweight. The 780G is truly the greatest pump ever, IMO. I look forward to trying it again someday when Medtronic releases that new CGM.

People should not avoid the 780G based on my experience alone - which is atypical. Honestly, IF the Guardian G4 Sensors are working adequately for someone (and they do work for many people) - then this would be my choice for a pump in a second.

I found that (while not approved in the USA) - that wearing them on my stomach - improved their performance for me (then, only off only by 15 mg/dl - which would have been acceptable) - but unfortunately it did not correct its abilitity detect and give me warnings to my LOWS - which for me - was a deal breaker.

Regards,

Doug
 
  • Hug
Reactions: Antje77 and Prancy

Prancy

Well-Known Member
Messages
383
Type of diabetes
Type 1
Treatment type
Pump
I can understand your decision Mr. Freeze. I share your concern over the G4, but they do normally work for me really well, though I’ve had them end on day six, not seven. That’s within their policy as acceptable. And, extending them for a second week, which I used to do with G3, isn’t feasible. They won’t go past 2-3 more days, so I don’t bother anymore. Just aim for 7 days and start with a fresh one. With G3 and my Medtronic 670, I could get almost 28 days on one sensor with accurate results, though I don’t advise anyone to do that. It is not recognized as acceptable by the company so no one do it.

I have the best of luck with my sensor on my arms. Even with my Dexcom, that was the case. The torso doesn’t work well for me.

Best of luck with your next treatmemt. i’ll look forward to seeing your progress.
 
  • Informative
Reactions: Antje77

Nicola M

Moderator
Staff Member
Moderator
Messages
820
Type of diabetes
Type 1
Treatment type
Pump
It's strange seeing how varied people's experiences can be with the same sensors. I've personally always found mine to be pretty accurate, although yes there have been times where it has been out for the most part it's been accurate. Funnily enough, Guardian sensors are approved for the abdomen over here in the UK which is where I do place mine. When compared to the Freestyle Libre 2 I've had a much better experience with Guardian which is probably helped more by the fact you can calibrate the sensors. It's always interesting to me to see people using insulins that aren't "approved" for use in the pump. I currently use Lyumjev which is also not approved but works perfectly fine.